Message No.

Error message

Trouble

Remedy

FAIL_19

LENS ERROR

Focus operation is failed.

Check the lens and lens cable. Check the
SY board. Replace the SY board if it is
damaged.

FAIL_21

ADJUSTMENT ERROR

The Z-axis operation is failed.

Check the lamp house-supplied motor and
harness. Replace the motor if it is dam-
aged. Check the SY board. Replace the
SY board if it is damaged.

FAIL_25

DEVICE ERROR

FPGA confi guration fails.

Turn off the power once, then turn on it
again.

FAIL_26

DEVICE ERROR

I

2

C communication error.

Turn off the power once, then turn on it
again.

FAIL_27

DEVICE ERROR

LVDS error

Turn off the power once, then turn on it
again.

FAIL_28

DEVICE ERROR

PLL error

Turn off the power once, then turn on it
again.

FAIL_29

DEVICE ERROR

Parity error

Turn off the power once, then turn on it
again.

FAIL_30

BALLAST ERROR

The power unit for lamp is interlocked.

Turn off the power once, then turn on it
again.

FAIL_36

DEVICE ERROR

Setting error of DVI.

Check the setting and set the operation
again.